A soon as I put the meter in 4.01 buffer solution WRNG appear on the screen and it remain so. When I press on O
I see OK1. I want to have OK2. I will try some other PH buffer manufacturer.
Are you sure the buffers did not get cross contaminated? Usually 4.01 will be out due to "buffering malpractice". Don't think is a brand problem, but who knows...
The Adwa meters have an option of 1 and/or 2 point calibration points, in which the second can be 4.01 or 10.01 (auto detected). But it has a second buffer set of 4.01/6.86/9.18.
Check if this last on is not set to default. Buffer 7.01 to 6.86 is not that far off and could trick the calibration and skew the curve to not accept 4.01. But my guess would first be buffer contam. Oh, by the way, is it new? Or have you been using it ok for some time now?
In any case, with the single 7.01 cal you should be in the ballpark between 5.5 and 6.5. Further from that is not to be trusted for measuring, but will be a clear indication nonetheless.
Also, have a cup of plan water handy. Rinse the probe and wipe the exterior. Calibrate 7.01. rinse, shake a little to clear the probe heads, and wipe the outside. Don't ever dip straight into the 4.01 of the buffers will contaminate!! (If you've ever done this, buy a new little 4.01 buffer sachet!!)
Put in 4.01 to calibrate. The thing is you have 12 seconds between buffers or it will error out. So once you OK 1 on 7.01, rinse, shake, wipe and dip!
Shake the probe inside the buffer solution a bit, a quick jiggle, so that no air bubbles/left overs swings the pH.
Also, leave a few drops of pH 7.01 in the cap of the pH meter. When storing, rinse, shake, wipe and cap it. This will ensure a longer lasting precision.
Also, I think Robwars should write a song called "Rise, Shake, Wipe and BOM!" as the soundtrack for his next featured presentation